ASTM E2119-20 - Standard Practice for Quality Systems for Conducting In Situ Measurements of Lead Content in Paint or Other Coatings Using Field-Portable X-Ray Fluorescence (XRF) Devices
September 1, 2020 - ASTM International

5.1 This practice provides procedures to generate and document QC data for ensuring that an XRF is operating within acceptable tolerances throughout the testing period when being used to collect lead results during a lead-based paint (LBP) inspection for the purposes of...

ASTM E2120-10(2016) - Standard Practice for Performance Evaluation of the Portable X-Ray Fluorescence Spectrometer for the Measurement of Lead in Paint Films
March 1, 2016 - ASTM International

5.1 The XRF instrument is used to measure the lead content in paint films in buildings and related structures in order to determine the potential lead hazard and the possible need for in-place control or abatement, or both. 5.2 This practice also is...

ASTM E2119 - Standard Practice for Quality Systems for Conducting In Situ Measurements of Lead Content in Paint or Other Coatings Using Field-Portable X-Ray Fluorescence (XRF) Devices
September 1, 2020 - ASTM

This practice covers the collection and documentation of quality control (QC) measurements for determining acceptable levels of instrumental performance when using field-portable energy-dispersive X-ray fluorescence spectrometry devices (XRFs) for the purposes of generating lead...

ASTM E2119-16 - Standard Practice for Quality Systems for Conducting In Situ Measurements of Lead Content in Paint or Other Coatings Using Field-Portable X-Ray Fluorescence (XRF) Devices
March 1, 2016 - ASTM International

5.1 This practice provides procedures to generate and document QC data for ensuring that an XRF is operating within acceptable tolerances throughout the testing period when being used to collect lead results during a lead-based pain (LBP) inspection for the purposes of generating...

ASTM PS88 - Provisional Standard Practice for the Determination of Lead in Paint, Settled Dust, Soil and Air Particulate by Field-Portable Electoanalysis
February 18, 1997 - ASTM

1. Scope 1.1 This provisional practice covers the analysis of extracts of environmental samples for lead content using field-portable electroanalytical devices. 1.2 Matrices of concern in this provisional practice include paint, settled dust, soil, and air particulate. 1.3...

ASTM E1979-21 - Standard Practice for Ultrasonic Extraction of Paint, Dust, Soil, and Air Samples for Subsequent Determination of Lead
November 1, 2021 - ASTM International

1.1 This practice covers an ultrasonic extraction procedure for the extraction of lead from environmental samples of interest in lead abatement and renovation (or related) work, for analytical purposes. 1.2 Environmental matrices of concern include dry paint films,...

ASTM E1644-21 - Standard Practice for Hot Plate Digestion of Dust Wipe Samples for the Determination of Lead
September 1, 2021 - ASTM International

5.1 This practice is intended for the digestion of lead in dust wipe samples collected during various lead hazard activities performed in and around buildings and related structures. 5.2 This practice is also intended for the digestion of lead in dust wipe...

ASTM E2914/E2914M-21 - Standard Practice for Ultrasonic Extraction of Lead from Composited Wipe Samples
September 1, 2021 - ASTM International

5.1 This practice is for use in the preparation of no more than four wipe samples collected from equally-sized areas in the same space combined to form a composited sample for subsequent determination of lead content. 5.2 This practice assumes use of wipes that meet...

ASTM E2913/E2913M-21 - Standard Practice for Hotplate Digestion of Lead from Composited Wipe Samples
September 1, 2021 - ASTM International

5.1 This practice is for use in the preparation of no more than four wipe samples combined to form a composited sample for subsequent determination of lead content. 5.2 This practice assumes use of wipes that meet Specification E1792 and should not be used unless the wipes meet...

ASTM E2120-00 - Standard Practice for the Performance Evaluation of the Portable X-Ray Fluorescence Spectrometer for the Measurement of Lead in Paint Films (Withdrawn 2009)
November 10, 2000 - ASTM International

The XRF instrument is used to measure the lead content in paint films in buildings and related structures in order to determine the potential lead hazard and the possible need for in-place control or abatement, or both. This practice also is to be...

ASTM PS116 - Provisional Practice for the Performance Evaluation of the Portable X-Ray Fluorescence Spectrometer for the Measurement of Lead in Paint Films
April 14, 1999 - ASTM

1. Scope 1.1 This provisional practice covers portable X-ray fluorescence (XRF) instruments intended for the measurement of lead in paint. It is intended that manufacturers apply the provisional practice to one unit of a particular model of an instrument when that model is...

ASTM E3074/E3074M-20 - Standard Practice for Clearance Examinations Following Lead Hazard Reduction Activities in Single Family Dwellings, in Individual Units of Multifamily Dwellings, and in Other Child-Occupied Facilities
September 1, 2020 - ASTM International

1.1 This practice covers visual assessment for the presence of deteriorated paint, surface dust, painted debris, and paint chips with environmental sampling of surface dust to determine whether a lead hazard exists at the time of sample collection, following...

ASTM E1775 - Standard Guide for Evaluating Performance of On-Site Extraction and Field-Portable Electrochemical or Spectrophotometric Analysis for Lead
September 1, 2020 - ASTM

This guide provides guidelines for determining the performance of field-portable quantitative lead analysis instruments. This guide applies to field-portable electroanalytical and spectrophotometric (including reflectance and colorimetric) analyzers. Sample matrices of concern herein include...

ASTM E1775-20 - Standard Guide for Evaluating Performance of On-Site Extraction and Field-Portable Electrochemical or Spectrophotometric Analysis for Lead
September 1, 2020 - ASTM International

4.1 This guide is intended for use in evaluating the performance of field-portable electroanalytical or spectrophotometric devices for lead determination, or both. 4.2 Desired performance criteria for field-based extraction procedures are provided. 4.3 Performance parameters of...

ASTM E1864-02 - Standard Practice for Evaluating Quality Systems of Organizations Conducting Facility and Hazard Assessments for Lead in Paint, Dust, Airborne Particulate, and Soil in and around Buildings and Related Structures (Withdrawn 2011)
July 10, 2002 - ASTM International

This practice provides the basic criteria to be used by accreditation bodies and others in evaluating the quality system, qualifications, and competency of organizations engaged in conducting facility and hazard assessments to determine the presence of lead contamination...

UNE-ISO/IEC TR 17026 IN - Conformity assessment -- Example of a certification scheme for tangible products.
December 5, 2017 - AENOR
A description is not available for this item.
ASTM E1864 - Standard Practice for Evaluating Quality Systems of Organizations Conducting Facility and Hazard Assessments for Lead in Paint, Dust, Airborne Particulate, and Soil in and Around Buildings and Related Structures
July 10, 2002 - ASTM

This practice covers the quality system requirements, competency and qualifications, to include minimum requirements for training, sampling design, personnel and equipment, duties, responsibilities, and services of organizations engaged in providing lead assessment services to...